fenair
16-09-2012, 05:11 PM
Hi All
Precision lpu150 freezer, Shop owner bought it second hand,
got it home and would not pull down, asked me to have a look,
I degassed, vac'd and pressure tested with n2, charged with 100gr R134a.
Pulls down to zero reasonably efficient, then steady progress from there.
Only ever managed to pull -19 once, this apparently took until the next day.
Once door has been opened a few times, it can't manage much around -10, and if left begins to warm up.
So i have had it over the week end, and triple vac'd it pressure tested for two hours and not a tad of movement on the gauge.
Re-gassed and same thing, it gets to zero OK, then struggles much past zero, puts up a alarm.
Gauge readings are +37 deg HP and -25 deg C LP.
Condenser is warm, pot is hot, and suction at outlet is cold, and ambient at the pot.
The evaporation is via the shelving consisting of three sections, lead in capillary is the top one this is frosted, the other two are not, but cold not iced though.
People i am thinking compressor is not circulating enough, but then why does it pull down up to freezing point, if the capillary was blocked then surely it would have no cooling effect?
Drier is warm, so i assume its not saturated, HP reading taken from the stub off the drier, and i have put a service T in the suction line back to the comp so i could vac either side of the compressor.
What have i missed??
Any thoughts would be appreciated..
Don't want to suggest a new comp and its something less expensive...
Regards and thanks in advance
Pat
Regards
Pat
Precision lpu150 freezer, Shop owner bought it second hand,
got it home and would not pull down, asked me to have a look,
I degassed, vac'd and pressure tested with n2, charged with 100gr R134a.
Pulls down to zero reasonably efficient, then steady progress from there.
Only ever managed to pull -19 once, this apparently took until the next day.
Once door has been opened a few times, it can't manage much around -10, and if left begins to warm up.
So i have had it over the week end, and triple vac'd it pressure tested for two hours and not a tad of movement on the gauge.
Re-gassed and same thing, it gets to zero OK, then struggles much past zero, puts up a alarm.
Gauge readings are +37 deg HP and -25 deg C LP.
Condenser is warm, pot is hot, and suction at outlet is cold, and ambient at the pot.
The evaporation is via the shelving consisting of three sections, lead in capillary is the top one this is frosted, the other two are not, but cold not iced though.
People i am thinking compressor is not circulating enough, but then why does it pull down up to freezing point, if the capillary was blocked then surely it would have no cooling effect?
Drier is warm, so i assume its not saturated, HP reading taken from the stub off the drier, and i have put a service T in the suction line back to the comp so i could vac either side of the compressor.
What have i missed??
Any thoughts would be appreciated..
Don't want to suggest a new comp and its something less expensive...
Regards and thanks in advance
Pat
Regards
Pat