    Ultra low Temp Refrigeration



    Hi there
    I am working on a Harris ultra low which has the 1st stage using 502 and 2nd stage using 503. I was wondering if anyone can tell me what pressures I would be looking at at start up on both stages and what pressures I would see when the box is down to temp.
    Also what would be good replacement refrigerants for ultra low refrigeration
    500
    502
    503

    Thanks for your help
    Eric



  2. #2
    Brian_UK's Avatar
    Brian_UK is offline Moderator I am starting to push the Mods: of RE Site Moderator : and general nice guy
    Join Date
    Mar 2001
    Location
    Dorset
    Age
    76
    Posts
    11,025
    Rep Power
    60

    Re: Ultra low Temp Refrigeration

    If you have a P/T chart for these refrigerants you should be able to find the relevant pressures when at ambient.

    The working pressures will be lower of course but you will get an indication.
    Brian - Newton Abbot, Devon, UK
    Retired March 2015

  3. #3
    Join Date
    Sep 2003
    Location
    ISRAEL
    Age
    72
    Posts
    4,248
    Rep Power
    46

    Re: Ultra low Temp Refrigeration

    Hi Eric,

    you can simply check the pressure and learn about the temp.
    its a cascade system.

    You can not use the new refrigerants unless you flush the systems and replace compressors.

    Chemi

  4. #4
    Join Date
    Dec 2008
    Location
    USA
    Age
    42
    Posts
    57
    Rep Power
    16

    Re: Ultra low Temp Refrigeration

    Harris doesn't have Thermo couples on their freezers so you can read temps on evaps and heat exchanger. Tapping into a system is not very well taken. Charges are very critical in UL applications

  5. #5
    Join Date
    Nov 2008
    Location
    australia
    Posts
    17
    Rep Power
    0

    Re: Ultra low Temp Refrigeration

    507, 404a high stage ,132a ,170 ,or 508 with propane

  6. #6
    jomama's Avatar
    jomama Guest

    Re: Ultra low Temp Refrigeration

    Before you do any service work on a ultra low cascade sys do a little research on low stage refrigerants 508b,503,114,13, ect all use r290 for oil return these are high psi refrigerants and can be dangerous if handle wrong. charging theses sys is completly differant than charging a standard sys. please ask questions for your safety and the safety of others.

Similar Threads

  1. Ultra low temp refrigeration
    By Nuklhed in forum Refrigeration Books
    Replies: 4
    Last Post: 25-03-2006, 03:05 PM
  2. sizing of low temp refrigeration cabinets
    By DALO in forum Technical Speculations
    Replies: 4
    Last Post: 15-01-2006, 08:31 PM
  3. Condensing temp in refrigeration
    By anup in forum Fundamentals
    Replies: 1
    Last Post: 02-10-2005, 11:18 AM
  4. Oil Migration - R22 medium temp refrigeration system
    By lurioda in forum Trouble Shooting
    Replies: 10
    Last Post: 09-03-2004, 03:38 PM

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •