12-01-2014, 06:36 PM #1
Neslab Merlin Chiller M33 problem after changing pump
Hello.
I've got a problem with Chiller M33. There was no water circulating, so I changed pump. After installing it, I had error code on the screen: "Er03" and after clearing it, appared : "ConF", what means BOM invalid (according to manual). Do you have any ideas, what is the reason of this situations? Type of pump is correct and only pump was changed. What should I do now?
Thanks for any ideas.
-
12-01-2014, 07:54 PM #2
Re: Neslab Merlin Chiller M33 problem after changing pump
http://www.mediafire.com/view/n6c1nk...ice_Manual.pdf
click on link above for manual- controller details.alarms further back in the manual
check water strainer, your conf alarm- check configuration of controller, otherwise its knackered
-
13-01-2014, 09:10 PM #3
-
13-01-2014, 09:21 PM #4
Re: Neslab Merlin Chiller M33 problem after changing pump
back of the manual states about settings for the controller- check they are all as per factory setting, check the display connections and other pcb connections havent been knocked and are seating correctly