Thread: Ultra Low temp too slow....
-
22-07-2013, 05:47 PM #1
Ultra Low temp too slow....
Hi,
I have recently been thrown into the cascade arena, and I am learning quickly. I have a problem with a Espec Thermal Shock Chamber.
The unit pulls down to -75, but takes a little to long, about an hour.. When I set the machine to cycle between -65 and 160, It cant keep up. It wont reach temp quick enough to allow any soak time. The charges have all been triple verified, so I am at a loss. Any suggestions?
Thx
-
22-07-2013, 08:40 PM #2
Re: Ultra Low temp too slow....
Well to give us a chance, running data could be useful...
1st stage suction pressure/discharge pressure/subcooling/superheat?
2nd stage suction pressure/discharge pressure/subcooling/superheat?
Ambient temp?
-
22-07-2013, 10:22 PM #3
Re: Ultra Low temp too slow....
How long did the unit take to pull down originally? Unless you know that you may find that the client is getting you to chase your tail.
Brian - Newton Abbot, Devon, UK
Retired March 2015
-
01-08-2013, 01:15 PM #4
Re: Ultra Low temp too slow....
these units take a long time to reach / stabilize tempreture and they run humidity to?? only one comp,viking?,,
look at specs, customer may be asking to much ,,haha
-
01-08-2013, 01:27 PM #5
Re: Ultra Low temp too slow....
Ya the colder you go the longer it takes, or the gear becomes really big! So less it is pulling more power then a third world nation I'd bank on an hour or more for pull down.
Now in Redvers Sask.
-
01-08-2013, 10:26 PM #6
Re: Ultra Low temp too slow....
OP seems to have disappeared.