    Ultra Low temp too slow....



    Hi,

    I have recently been thrown into the cascade arena, and I am learning quickly. I have a problem with a Espec Thermal Shock Chamber.

    The unit pulls down to -75, but takes a little to long, about an hour.. When I set the machine to cycle between -65 and 160, It cant keep up. It wont reach temp quick enough to allow any soak time. The charges have all been triple verified, so I am at a loss. Any suggestions?

    Thx



  2. #2
    Join Date
    Feb 2007
    Location
    UK
    Posts
    2,479
    Rep Power
    45

    Re: Ultra Low temp too slow....

    Well to give us a chance, running data could be useful...

    1st stage suction pressure/discharge pressure/subcooling/superheat?
    2nd stage suction pressure/discharge pressure/subcooling/superheat?

    Ambient temp?


  3. #3
    Brian_UK's Avatar
    Brian_UK is offline Moderator I am starting to push the Mods: of RE Site Moderator : and general nice guy
    Join Date
    Mar 2001
    Location
    Dorset
    Age
    76
    Posts
    11,192
    Rep Power
    60

    Re: Ultra Low temp too slow....

    How long did the unit take to pull down originally? Unless you know that you may find that the client is getting you to chase your tail.
    Brian - Newton Abbot, Devon, UK
    Retired March 2015

  4. #4
    Join Date
    Nov 2008
    Location
    australia
    Posts
    17
    Rep Power
    0

    Re: Ultra Low temp too slow....

    these units take a long time to reach / stabilize tempreture and they run humidity to?? only one comp,viking?,,
    look at specs, customer may be asking to much ,,haha

  5. #5
    Join Date
    Apr 2006
    Location
    Not so sunny coast (BC Canada)
    Age
    41
    Posts
    1,645
    Rep Power
    23

    Re: Ultra Low temp too slow....

    Ya the colder you go the longer it takes, or the gear becomes really big! So less it is pulling more power then a third world nation I'd bank on an hour or more for pull down.
    Now in Redvers Sask.

  6. #6
    Brian_UK's Avatar
    Brian_UK is offline Moderator I am starting to push the Mods: of RE Site Moderator : and general nice guy
    Join Date
    Mar 2001
    Location
    Dorset
    Age
    76
    Posts
    11,192
    Rep Power
    60

    Re: Ultra Low temp too slow....

    OP seems to have disappeared.