Results 1 to 3 of 3
  1. #1
    Join Date
    Feb 2008
    Location
    england
    Posts
    505
    Rep Power
    21

    probe calibration



    if i have a calibrated thermometer can i calibrate a thermostat and issue a calibration certificate



  2. #2
    Brian_UK's Avatar
    Brian_UK is offline Moderator I am starting to push the Mods: of RE Site Moderator : and general nice guy
    Join Date
    Mar 2001
    Location
    Dorset
    Age
    76
    Posts
    11,192
    Rep Power
    60

    Re: probe calibration

    I don't see why not, as long as you can provide an audit trail for the calibration.

    It may depend on the customer though.
    Brian - Newton Abbot, Devon, UK
    Retired March 2015

  3. #3
    Join Date
    Aug 2007
    Location
    Sydney
    Age
    51
    Posts
    1,094
    Rep Power
    23

    Re: probe calibration

    Hi Martin
    I think you could issue a certificate saying it has been checked against a calibrated thermometer but not a calibration certificate saying it has been calibrated (we could get the Chinese whispers effect). There are several different types of temp sensors which have varying response times, hysteresis, repeatability, changes with airflow rates etc. These measurements for calibration are carried out in lab conditions where temp, humidity, pressure etc are measured and compared under stringent conditions. Do you have a procedure for checking a sensor like number of measurements and times etc?