    Hello,

    I hope someone can help. I'm a scientist, working in a university, building my own experimental setup. I want to modify a Thermo Neslab RTE 7 Digital One heater/chiller (water bath recirculator).

    The recirculator measures the temperature in its own water bath, I want to move the temperature probe to another (external) water bath.

    After opening up the controller (top) part of the device I'm guessing that the temperature probe is a metal rod, partially encased in black plastic. It has 4 tiny wires running into it: 2 orange wires and 2 blue wires.

    First question: Have I correctly identified the probe?

    Second question: If so, can you forsee any problems with me removing the probe and extending the length of the wires from its current length of 1 foot to a new length of 6 feet, and placing it in an external water bath.

    I'll have to carefully program the controller to handle the new response time, but I hope that will be easy enough to do.

    Any advice at all is appreciated.

    Damien



  2. #2
    Brian_UK's Avatar
    Brian_UK is offline Moderator I am starting to push the Mods: of RE Site Moderator : and general nice guy
    Join Date
    Mar 2001
    Location
    Dorset
    Age
    76
    Posts
    11,025
    Rep Power
    60

    Re: modifying a Thermo Neslab RTE 7 Digital One heater/chiller

    Hi Damien and welcome to the forum.

    Any clue as to the make of the controller?

    Wiki has this info on 4-wire sensors...
    http://en.wikipedia.org/wiki/Four-terminal_sensing

    or this is better..
    http://en.wikipedia.org/wiki/Resistance_thermometer
    Last edited by Brian_UK; 01-02-2012 at 09:07 PM.
    Brian - Newton Abbot, Devon, UK
    Retired March 2015

  3. #3
    Join Date
    Jul 2006
    Location
    USA
    Age
    73
    Posts
    865
    Rep Power
    26

    Re: modifying a Thermo Neslab RTE 7 Digital One heater/chiller

    You can find some information on that chiller here.
    http://www.coleparmer.com/Assets/man...3500Series.pdf

    Although I do not see your chiller listed this is a good resource.
    http://www.chillercity.com/manuals.php

  4. #4
    Join Date
    Feb 2012
    Location
    United States
    Posts
    2
    Rep Power
    0

    Re: modifying a Thermo Neslab RTE 7 Digital One heater/chiller

    Thanks for the feedback. I have looked at the manuals for the device, but they did not contain the kind of detail I required. By "controller" I mean the "Digital One" top part of the device (includes heater and a bunch of electronics).

    But it's going ok so far: I managed to to remove the RTD probe from the unit without problems. Looks like I simply need to extend the 4 cables cable by ~6 ft by inserting a new joint. I also have a second option of buying a second probe with sufficiently long cables. Chris at chillercity.com was really helpful. I should be OK from here on in.

    Damien

  5. #5
    Join Date
    Dec 2011
    Location
    Ireland
    Posts
    44
    Rep Power
    0

    Re: modifying a Thermo Neslab RTE 7 Digital One heater/chiller

    Sounds like you've found your probe(rtd1). I can't see why extending it won't work, but failing that, a second external probe (rtd2)is available for those chillers I think. When circulating to an external bath, remember when sizing the bath to keep within the cooling/heating capacity range of the chiller.
    Best of luck,
    Paul.

Posting Permissions