Results 1 to 11 of 11
    Oct 2003
    56
    5,519
    28
    Testo 560, apparently fault
    In pdf you will find pressure/temperature scale of water and besides of this table, 3 pictures of the Testo560, 2 in Pa and 1 in Microns.
    You will see that the figures on the screen of the Testo 560 doesn't match with those in the Coolpack table.
    How it comes? I can't find what's wrong.
    Attached Files Attached Files
    Mar 2006
    Merate (LC) - Italy
    47
    2,553
    Re: Testo 560, apparently fault
    I think you are confusing the vapour pressure with the saturated evaporating pressure.
    The triple point of water is 0C, so vapor equlibrium only exists above 0C.
    Oct 2003
    56
    5,519
    28
    The Testo is in vacuum mode and you can see in the right corner H2O
    Mar 2006
    Merate (LC) - Italy
    47
    2,553
    Resolution of instrument is 80microns and accuracy is +/-0.5% of scale. I don't think it can do any better than that. Try zeroing the instrument.
    Jul 2008
    Quote Originally Posted by NoNickName View Post
    Resolution of instrument is....
    I think that accuracy (or resolution) has nothing to do to it.
    The conversion from pressure to temperature goes wrong.
    One of the two is at least wrong, but who (Testo or Coolpack?)
    CoolPack is freeware and you can find it on google
    Mar 2006
    Merate (LC) - Italy
    47
    2,553
    Coolpack can't be wrong. It's a mathematical software. I cross confirmed the output values with another software.
    Oct 2003
    56
    5,519
    28
    NoNickname, Gerrit has a point I think, even with the given inaccuracy, the conversion between pressure and temperature must match anyhow because the Testo measures pressure and then converts it via the software to a temperature.
    figaro's Avatar
    figaro Guest
    I bet it has to do with the sea level adjustment in the testo. Its adjustable to where you reside.
    Oct 2003
    56
    5,519
    28
    It starts from zero and the adjustment was done. Still not solved.
    Oct 2003
    56
    5,519
    28
    N...the conversion between pressure and temperature must match anyhow because the Testo measures pressure and then converts it via the software to a temperature.
    I meant, the Testo is converting its pressure to a saturated evaporating temperature because that's table the Testo uses, the saturated one.
    Jul 2008
    Hey Peter_1, and Figaro
    This is my Testo560, and he is set to sea level
    Until now no reply received from Testo
    Still in limbo
    I meet them soon on exhibitions
  1. Hitachi "31" fault
    By ronaldo9 in forum Trouble Shooting
    Last Post: 22-02-2010, 07:26 PM
  2. Daikin r407c 2 pipe vrv ja fault code any offers?
    By Contactor in forum Air Conditioning
    Last Post: 28-07-2009, 01:26 AM
  3. Fault with Tango cassette
    By airefresco in forum Air Conditioning
    Last Post: 19-09-2007, 08:42 AM
  4. UA Daikin Fault Code
    By jedi in forum Air Conditioning
    Last Post: 23-07-2007, 01:44 PM
  5. Chiller fault
    By iceman007 in forum Trouble Shooting
    Last Post: 27-06-2006, 06:15 AM