    Argweld purge monitor's sensor failure



    Hello.
    I have seen two or three Argweld purge monitors MKIII which would not want get calibrated - they show values from some minus 5 to plus 5 but never 20.9 (the usual oxygen %). Evidently, faulty sensors, but what made them out of order? Hot gas got on them? (They were used during welding of stainless pipes).



  2. #2
    Brian_UK's Avatar
    Brian_UK is offline Moderator I am starting to push the Mods: of RE Site Moderator : and general nice guy
    Join Date
    Mar 2001
    Location
    Dorset
    Age
    76
    Posts
    11,025
    Rep Power
    60

    Re: Argweld purge monitor's sensor failure

    Any of the following happened or done/not done? (From tech docs.
    +
    (MkIII) The benefit of this is the 18 - 36 months life of the new sensor.

    Sensor life is reduced by continuous exposure to high levels of oxygen.

    Life will be greater if the monitor is used frequently, that is, in measuring weld purging by removing oxygen from the system.


    If the monitor is unused for long periods, sensor life can be increased by sealing the sensor or storing the instrument in an argon filled plastic bag or container.
    +

    Brian - Newton Abbot, Devon, UK
    Retired March 2015

  3. #3
    Join Date
    Aug 2008
    Location
    Baltics
    Age
    62
    Posts
    786
    Rep Power
    16

    Re: Argweld purge monitor's sensor failure

    Thank you! I have passed this info to the person whom those tools belong.