I stupidly adjusted the calibration on my thermometer when it was actually down to a faulty probe.
I'm now trying to re calibrate it myself and used boiling water as i know that it's 100c.

Is this going to work?

:confused:

Toosh
21-04-2010, 10:09 PM
I stupidly adjusted the calibration on my thermometer when it was actually down to a faulty probe.
I'm now trying to re calibrate it myself and used boiling water as i know that it's 100c.

Is this going to work?

:confused:

Hi usually to calibrate, use ice and water, fusion of ice theory the water will be 0 degrees

sedgy
21-04-2010, 10:21 PM
I thought electronic tools had to be calibrated and taged by an aproved firm every 2 or 3 years?

Brian_UK
21-04-2010, 10:50 PM
Only if your customer demands it.

Gingerair
21-04-2010, 11:03 PM
Some firms only check the calibration against accepable values then issue a pass or fail cert without attempting to carry out any adjustments.

Was told by someone at the RS calibration centre that they test all instruments using the same procedure & equipment, but charge extra just for supplying the result on a UKAS cert...

ref717
24-04-2010, 05:09 PM
What are the temp. ranges of your electronic thermometer?,you just cannot calibrate by yourself if you have a wide temp. range like -50°C to 400°C, unless you have an acceptable temp. controller that is calibrated by an approved firm.

sedgy
24-04-2010, 08:23 PM
only if the customer requires it?
I would have thought any tradesman would want it correctly calibrated ?

Brian_UK
24-04-2010, 10:16 PM
I thought electronic tools had to be calibrated and taged by an aproved firm every 2 or 3 years?


only if the customer requires it?
I would have thought any tradesman would want it correctly calibrated ?You weren't commenting on that aspect were you?

sedgy
24-04-2010, 11:21 PM
if you say not , tell me what was I COMMENTING about? if you know, tell us all you seem to know everything,?

deepee
25-04-2010, 05:36 AM
what is the probe type in your thermometer ? You can calibrate your RTD / Thermocouple indicators with calibrators.

Peter_1
25-04-2010, 11:52 AM
We have within our clients one of the biggest calibration companies in the Benelux.
All the rooms are maintained within +/-0.2°C.
They calibrate temperature measuring devices in some sort of oil heated to 250°C and melting ice made of pure water.
If you have two points, then your device will perform very precise in the ranges we usually work, even with boiling water as the upper limit.